Objectives
This tutorial aims to familiarize you with the
following principles of weighing:
Choosing
the Proper Instrument for Weighing
All weighing in the chemistry laboratory is done
with a balance. However, depending on the requirements of the experiment
at hand, different balances may be used.
Preparative work, such as measuring out a rough
amount of an excess reagent, does not require a high level of accuracy.
For example, if an experimental procedure calls for 1.0 grams of
a reagent known to be in excess in the reaction, amounts of 0.9
or 1.1 grams are perfectly acceptable. For these purposes, a top-loading
balance is used to obtain an initial, approximate mass that is within
the range specified by the procedure. The sample's mass is then
determined more accurately before being presented in a laboratory
report. A top-loading balance is a relatively simple electronic
device that measures the mass of an object to an uncertainty of
±0.05 grams, thus being suitable for preparative applications
in the chemistry laboratory.
In analytical work, such as precisely determining
the mass of a lead salt precipitate meticulously isolated from an
aqueous solution, an analytical balance should be used to minimize
uncertainty. An analytical balance is a highly sensitive instrument
that is much more prone to errors from environmental conditions
and requires much more careful operation than a top-loading balance.
This balance measures an object's mass to an uncertainty of ±0.00005
grams and is used any time a very accurate mass determination is
needed.
Question 1
Determine whether each of the tasks below is best
accomplished with a top-loading or an analytical balance.
- Measuring the mass of a lead salt precipitate to the nearest
0.1 gram.
- Weighing out anhydrous solid copper sulfate for a reaction
where the procedure calls for 1.5 grams of this chemical.
- Determining the water content of hydrated copper sulfate by
weighing a sample of this salt prior to and following drying in
an oven.
- Weighing out sodium oxalate for a solution that will be used
to standardize a solution of potassium permanganate via a Fowler
and Bright modified titration.
Technique
Notes for More Accurate Weighing
Presented below are some general pointers to increase
the accuracy of your mass measurements, no matter which balance
you use.
| Weigh objects at room temperature. |
A warm or hot object will create a convection
current in the air around the balance pan. This fluctuating
force reduces the air pressure on the balance pan and can make
it difficult to obtain a stable reading. |
| Weigh only dry objects. |
Moisture can corrode the balance pan. Moisture
evaporating from your sample can lead to unstable mass readings. |
| Use approximate amounts and then measure them
accurately. |
Approximate amounts should be weighed out
on a top-loading balance and then weighed accurately on an
analytical balance.
For example, if the lab manual asks you to
weigh out 0.8 grams of a material, a range of 0.7-0.9 grams
as measured by an analytical balance is acceptable for the
reaction. It would take much too long to weigh out exactly
0.8000 grams of the material. Instead, weigh your approximately
measured out sample on an analytical balance to obtain its
accurate mass. This mass, accurate to the nearest 0.1 mg,
can be stated in your laboratory reports. |
| Use the same balance. |
It is critical to always use the same balance
especially if calculations require more than one mass measurement,
as each device might be calibrated slightly differently. |
| Use weigh boats. |
Weigh boats are containers used to prevent
reagents from contacting the balance pan. They are made of polypropylene,
a plastic that does not adsorb water. They are inexpensive and
do not need to be handled with care. If one is torn or too dirty
to be wiped clean, simply discard it. |
Using
the Top-Loading Balance
Provided below is an interactive Macromedia Flash
walk-through that shows you the steps in properly using the top-loading
balance.
Using
the Analytical Balance
Analytical balances are highly sensitive instruments
and provide very accurate mass measurements. However, this accuracy
comes at a price: not only are they expensive, but analytical balances
are also quite susceptible to environmental conditions that can
affect the mass reading. The table below outlines several events
that can cause the mass reading to shift by several ten-thousandths
of a gram or be unstable. The remedies for each problem are also
shown below.
| Source of Error |
Remedy for Error |
| The floor of the balance room shaking or tilting very slightly
as people move around the room. |
Place the balance on a heavy table, near an outside wall where
floor motions will be minimized. |
| Leaning on or laying a hand onto the balance table. |
Place the balance on a heavy table and try to minimize the
pressure applied to the table. |
| Air currents lifting the balance pan. |
Close the sliding doors on the balance. |
| Wet samples becoming lighter as the moisture they contain
evaporates off. |
Make sure that the sample is dry before it is weighed. |
| Ultra-dry samples becoming heavier as they adsorb moisture
from the atmosphere. |
Let the sample sit at room temperature for five minutes prior
to weighing it. |
| Weighing warm objects creates a convection current that lifts
the balance pan. |
Let the sample sit at room temperature for five minutes prior
to weighing it. |
| Handling glass objects with bare fingers deposits oil on them
and increases their mass. |
Hold object through a strip of paper or wear latex gloves. |
| Breathing on a glass object condenses moisture on it and increases
its mass. |
Don't breathe on the object. |
The most accurate reading on an analytical balance
will be the first stable weight displayed. The most accurate reading
is not necessarily the value that seems to persist for a longer
time.
The interactive Macromedia Flash walk-through provided
below shows you the steps in properly using the analytical balance.
Weighing
by Differences
Weighing by differences is a technique that is
used when it is important to know the precise amount of a sample
that has been delivered into a reaction mixture. For this
purpose, It is insufficient to simply tare a weigh boat and accurately
weigh the sample that has been transferred into it, because some
of the sample inevitably sticks to the weigh boat and does not make
it into the reaction beaker on a subsequent transfer. Weighing by
differences tells you exactly how much of the sample made it out
of the weigh boat.
The interactive Macromedia Flash walk-through provided
below shows you the steps for weighing a sample by differences.
Question 2
You are weighing out a sample of copper sulfate
by differences. You obtain a weigh boat, place it onto a top-loading
balance, tare the balance with the empty weigh boat, and measure
out your sample. Satisfied that your sample's weight on the top-loading
balance is within the range specified by the lab manual, you complete
the weighing by differences on an analytical balance and transfer
a precisely measured amount of the copper sulfate into a reaction
beaker.
Your lab partner comes up to you and informs you
that you accidentally did your weighing with her used weigh boat,
which contained small amounts of copper sulfate. How does this affect
your result? Do you need to repeat your measurement?
Summary
This tutorial on weighing objects has presented
the following topics:
- How to choose between a top-loading and an analytical balance.
- Technical pointers for more accurate weighing.
- Correct technique for weighing with a top-loading balance.
- Correct technique for weighing with an analytical balance.
- The advantages of and correct technique for weighing by differences.
|